I'm trying to get an LIS3DH accelerometer working and I have a question about the output data.
It seems to be very ''noisy''. I have the board with the accelerometer on a flat table.
I've set the data rate to 100hz and 2g fullscale , high res mode on.
The data seems to jump all over the place. If I tilt the board fully one direction I get about 16000 counts
and fully the other direction I get about -16000 counts, which seems to be about right.Still very noisy
around the main reading.
when it is flat on the table I get anywhere from +128 to - 128 counts. Counts change in increments of 16.
Is this normal behavior?
Cheers
Neddie
2016-11-02 08:41 AM
The output data are 12bit, left justified.
You have to divide the raw data by 16 (shift right by 4). For 2g range the sensitivity is 1mg/LSB so you will get directly the value in mg.
The noise of +-128 counts means +-8mg which is in line with the datasheet.
Best regards.
Miroslav