We are currently evaluating the L6470 to drive analytic high-pressure pumps for Mass spectrometry.The final goal is to drive a stepper-motor in high precision micro-stepping mode at around 1.7 Nm at a maximum Speed of around 100 rpm.For evaluation we used a small bipolar stepper-motor from a previous test with the following parameters:As motor driver we used the EVAL6470H board attached to a STEVAL-PCC009V1 USB-SPI Interface in combination with the dSPIN Evaluation Tool 1.7.
Further Parameters:Here our Questions:
We really looking forward for your answers and thank you for your efforts in advance!
Sincerely yours #stepper-motor-driver #stepper-motor #l64702012-11-14 08:08 AM
Hello,
The stall issue is probably related to a resonance condition. When the motor is unloded the mid-point resonance are heavily stimulated, in many cases adding a small load at the motor shaft dramatically reduce the resonances allowing the motor to run correctly.You can also check the effectiveness of the BEMF compensation setup. You should consider that the BEMF compensation tool is based on an approximated model of the stepper motor, so it could not returns the optimal configuration.You can read the following AN for further details about the optimization of the voltage mode driving algorithm:http://www.st.com/internet/com/TECHNICAL_RESOURCES/TECHNICAL_LITERATURE/APPLICATION_NOTE/DM00061093.pdfThe communication error issue should be related to an attempt to write some configuration registers when it is not allowed (motor running or bridges enabled). Please consider that the configuration tool always write all the configuration register, so in order to avoid the communication error you need to switch the device in high impedance status.RegardsEnrico