SPI Signal Shape Issue

Hello,

I am try to control an external DAC system with the SPI interface. The clock signal shape looks like the attached file. I didn't figure out the meaning of the signal. Yellow is the clock signal and the blue is CS signal.0693W00000aJE10QAG.jpg

This topic has been closed for replies.
AScha.3
AScha.3Best answer
Super User
March 17, 2023

yes. Need to change to something useful. set -> button Aquire-> mode:normal, mem-depth: auto , anti-alias:off ; + buton Display -> type: vector .

+

set your probe to x10 , an in dso probe menu also x10 setting.

(and it was NOT default setting ! )

if want some gaming: just try -- press button AUTO , with signal active; scope tries to set it then automatic good.

If you feel a post has answered your question, please click on " Best Answer ".

24 replies

AE104
AE104Author
Associate III
March 17, 2023

..

AScha.3
Super User
March 17, 2023

better.

probe still at x1 ! you cannot test cmos connections at > 10MHz with x1 probe setting.

set x10 at probe switch and in scope menu.

If you feel a post has answered your question, please click on " Best Answer ".
AE104
AE104Author
Associate III
March 17, 2023

I set the 10x both the probe and settings in the oscilloscope.

0693W00000aJElwQAG.jpg0693W00000aJEm1QAG.jpg

AScha.3
Super User
March 17, 2023

x10 probe setting needs adjustment - did you adjust ?

+ BW limit 20 M is set.

do you make some game here ? who is the biggest idiot ? or what ?

sorry.

want help - or play game ?

If you feel a post has answered your question, please click on " Best Answer ".
AE104
AE104Author
Associate III
March 17, 2023

No game. What adjustment do I need?

AScha.3
Super User
March 17, 2023

bw limit off.

and adjust probes - i hope you know how to do.

If you feel a post has answered your question, please click on " Best Answer ".
AE104
AE104Author
Associate III
March 17, 2023

Yes I did it.

AE104
AE104Author
Associate III
March 17, 2023

This is the latest measurement that I have.

AScha.3
Super User
March 18, 2023

again sample rate much too low. 500 ks - how you do this ?

+

or do you make "SINGLE" shot and then turn around the timing? this could make such pictures.

show a measurement without any change or "adjustment" after the single shot.

If you feel a post has answered your question, please click on " Best Answer ".
AE104
AE104Author
Associate III
March 29, 2023

Sorry for late reply. Honestly, I don't know, for example now the sampling rate of the oscilloscope is 1 GSa/s.

You're right. I captured only this zoom in figure. Before zoomed in, it was square wave.

AScha.3
Super User
March 21, 2023

When your question is answered, please mark this topic as answered by choosing Select as Best for the preferred answer, as also explained here. This will help other users find that answer faster.