I am trying to measure the XYZ values of the sensor AIS328DQ. I am using 100K I2C master to read the values from the sensor.
I am not getting any measurements that are consistent. Below are the configurations and readings for XYZ from the sensor. Note that the values of XYZ are fluctuating.
Note that the below test measurements configures the registers 20, 21, 22, 23, 24 with values 3F, 00, 3F, 31, 00 respectively.
The X Y Z values are read from registers (28, 29), (2A, 2B), (2C, 2D) respectively, also note that the readings for XYZ are read multiple times for our experiments.
------------- Below is the configuration and measurement log ---------
Export Time: 2015-03-12 14:57:41
Port 0
Adapter HW_Version: 3.00 FW_Version: 3.50
"Time","Module","Read/Write","Master/Slave","Features","Bitrate","Address","Length","Data"
"2015-03-12 14:57:30.803","","","","","","","","Configure: I2C=1 SPI=1 GPIO=0"
"2015-03-12 14:57:30.803","","","","","","","","Power Control Enabled"
"2015-03-12 14:57:30.805","I2C","","","","","","","I2C Pullups Disabled"
"2015-03-12 14:57:31.309","","","","","","","","Signal Level Set to: Failed -6"
"2015-03-12 14:57:31.322","I2C","","","","","","","I2C Bitrate Set to: 100"
"2015-03-12 14:57:31.326","I2C","W","M","--S","100","0x18","1","0F"
"2015-03-12 14:57:31.331","I2C","R","M","---","100","0x18","2","32 32"
"2015-03-12 14:57:31.335","I2C","W","M","--S","100","0x18","2","20 3F"
"2015-03-12 14:57:31.339","I2C","W","M","--S","100","0x18","2","21 00"
"2015-03-12 14:57:31.344","I2C","W","M","--S","100","0x18","2","22 3F"
"2015-03-12 14:57:31.348","I2C","W","M","--S","100","0x18","2","23 31"
"2015-03-12 14:57:31.352","I2C","W","M","--S","100","0x18","2","24 00"
"2015-03-12 14:57:31.355","I2C","W","M","--S","100","0x18","1","20"
"2015-03-12 14:57:31.358","I2C","R","M","---","100","0x18","2","3F 3F"
"2015-03-12 14:57:31.361","I2C","W","M","--S","100","0x18","1","21"
"2015-03-12 14:57:31.365","I2C","R","M","---","100","0x18","2","00 00"
"2015-03-12 14:57:31.371","I2C","W","M","--S","100","0x18","1","22"
"2015-03-12 14:57:31.375","I2C","R","M","---","100","0x18","2","3F 3F"
"2015-03-12 14:57:31.380","I2C","W","M","--S","100","0x18","1","23"
"2015-03-12 14:57:31.384","I2C","R","M","---","100","0x18","2","31 31"
"2015-03-12 14:57:31.387","I2C","W","M","--S","100","0x18","1","24"
"2015-03-12 14:57:31.392","I2C","R","M","---","100","0x18","2","00 00"
"2015-03-12 14:57:31.397","I2C","W","M","--S","100","0x18","1","28"
"2015-03-12 14:57:31.401","I2C","R","M","---","100","0x18","9","00 00 00 00 00 00 00 00 00"
"2015-03-12 14:57:31.405","I2C","W","M","--S","100","0x18","1","29"
"2015-03-12 14:57:31.411","I2C","R","M","---","100","0x18","9","00 00 00 00 00 00 00 00 00"
"2015-03-12 14:57:31.526","I2C","W","M","--S","100","0x18","1","2A"
"2015-03-12 14:57:31.532","I2C","R","M","---","100","0x18","6","20 20 20 20 20 20"
"2015-03-12 14:57:31.536","I2C","W","M","--S","100","0x18","1","2B"
"2015-03-12 14:57:31.539","I2C","R","M","---","100","0x18","6","00 00 00 00 00 00"
"2015-03-12 14:57:31.667","I2C","W","M","--S","100","0x18","1","2C"
"2015-03-12 14:57:31.676","I2C","R","M","---","100","0x18","3","20 20 20"
"2015-03-12 14:57:31.684","I2C","W","M","--S","100","0x18","1","2D"
"2015-03-12 14:57:31.691","I2C","R","M","---","100","0x18","3","EF EF EF"
"2015-03-12 14:57:31.822","I2C","W","M","--S","100","0x18","1","28"
"2015-03-12 14:57:31.831","I2C","R","M","---","100","0x18","9","50 50 50 50 50 50 50 50 50"
"2015-03-12 14:57:31.838","I2C","W","M","--S","100","0x18","1","29"
"2015-03-12 14:57:31.842","I2C","R","M","---","100","0x18","9","00 00 00 00 00 00 00 00 00"
"2015-03-12 14:57:31.966","I2C","W","M","--S","100","0x18","1","2A"
"2015-03-12 14:57:31.973","I2C","R","M","---","100","0x18","6","40 40 40 40 40 40"
"2015-03-12 14:57:31.982","I2C","W","M","--S","100","0x18","1","2B"
"2015-03-12 14:57:31.990","I2C","R","M","---","100","0x18","6","00 00 00 00 00 00"
"2015-03-12 14:57:32.121","I2C","W","M","--S","100","0x18","1","2C"
"2015-03-12 14:57:32.130","I2C","R","M","---","100","0x18","3","40 40 40"
"2015-03-12 14:57:32.138","I2C","W","M","--S","100","0x18","1","2D"
"2015-03-12 14:57:32.147","I2C","R","M","---","100","0x18","3","F0 F0 F0"
"2015-03-12 14:57:32.276","I2C","W","M","--S","100","0x18","1","28"
"2015-03-12 14:57:32.279","I2C","R","M","---","100","0x18","9","10 10 10 10 10 10 10 10 10"
"2015-03-12 14:57:32.283","I2C","W","M","--S","100","0x18","1","29"
"2015-03-12 14:57:32.286","I2C","R","M","---","100","0x18","9","00 00 00 00 00 00 00 00 00"
"2015-03-12 14:57:32.403","I2C","W","M","--S","100","0x18","1","2A"
"2015-03-12 14:57:32.407","I2C","R","M","---","100","0x18","6","80 80 80 80 80 80"
"2015-03-12 14:57:32.413","I2C","W","M","--S","100","0x18","1","2B"
"2015-03-12 14:57:32.417","I2C","R","M","---","100","0x18","6","00 00 00 00 00 00"
"2015-03-12 14:57:32.542","I2C","W","M","--S","100","0x18","1","2C"
"2015-03-12 14:57:32.553","I2C","R","M","---","100","0x18","3","20 20 20"
"2015-03-12 14:57:32.559","I2C","W","M","--S","100","0x18","1","2D"
"2015-03-12 14:57:32.565","I2C","R","M","---","100","0x18","3","F0 F0 F0"
"2015-03-12 14:57:32.684","I2C","W","M","--S","100","0x18","1","28"
"2015-03-12 14:57:32.687","I2C","R","M","---","100","0x18","9","40 40 40 40 40 40 40 40 40"
"2015-03-12 14:57:32.690","I2C","W","M","--S","100","0x18","1","29"
"2015-03-12 14:57:32.694","I2C","R","M","---","100","0x18","9","00 00 00 00 00 00 00 00 00"
"2015-03-12 14:57:32.822","I2C","W","M","--S","100","0x18","1","2A"
"2015-03-12 14:57:32.828","I2C","R","M","---","100","0x18","6","70 70 70 70 70 70"
"2015-03-12 14:57:32.832","I2C","W","M","--S","100","0x18","1","2B"
"2015-03-12 14:57:32.836","I2C","R","M","---","100","0x18","6","00 00 00 00 00 00"
"2015-03-12 14:57:32.963","I2C","W","M","--S","100","0x18","1","2C"
"2015-03-12 14:57:32.969","I2C","R","M","---","100","0x18","3","B0 B0 B0"
"2015-03-12 14:57:32.973","I2C","W","M","--S","100","0x18","1","2D"
"2015-03-12 14:57:32.980","I2C","R","M","---","100","0x18","3","F0 F0 F0"
"2015-03-12 14:57:33.104","I2C","W","M","--S","100","0x18","1","28"
"2015-03-12 14:57:33.111","I2C","R","M","---","100","0x18","9","40 40 40 40 40 40 40 40 40"
"2015-03-12 14:57:33.121","I2C","W","M","--S","100","0x18","1","29"
"2015-03-12 14:57:33.128","I2C","R","M","---","100","0x18","9","00 00 00 00 00 00 00 00 00"
"2015-03-12 14:57:33.259","I2C","W","M","--S","100","0x18","1","2A"
"2015-03-12 14:57:33.269","I2C","R","M","---","100","0x18","6","B0 B0 B0 B0 B0 B0"
"2015-03-12 14:57:33.276","I2C","W","M","--S","100","0x18","1","2B"
"2015-03-12 14:57:33.285","I2C","R","M","---","100","0x18","6","00 00 00 00 00 00"
"2015-03-12 14:57:33.414","I2C","W","M","--S","100","0x18","1","2C"
"2015-03-12 14:57:33.423","I2C","R","M","---","100","0x18","3","50 50 50"
"2015-03-12 14:57:33.430","I2C","W","M","--S","100","0x18","1","2D"
"2015-03-12 14:57:33.437","I2C","R","M","---","100","0x18","3","F0 F0 F0"
"2015-03-12 14:57:33.571","I2C","W","M","--S","100","0x18","1","28"
"2015-03-12 14:57:33.573","I2C","R","M","---","100","0x18","9","20 20 20 20 20 20 20 20 20"
"2015-03-12 14:57:33.576","I2C","W","M","--S","100","0x18","1","29"
"2015-03-12 14:57:33.581","I2C","R","M","---","100","0x18","9","00 00 00 00 00 00 00 00 00"
"2015-03-12 14:57:33.696","I2C","W","M","--S","100","0x18","1","2A"
"2015-03-12 14:57:33.703","I2C","R","M","---","100","0x18","6","60 60 60 60 60 60"
"2015-03-12 14:57:33.710","I2C","W","M","--S","100","0x18","1","2B"
"2015-03-12 14:57:33.717","I2C","R","M","---","100","0x18","6","00 00 00 00 00 00"
"2015-03-12 14:57:33.835","I2C","W","M","--S","100","0x18","1","2C"
"2015-03-12 14:57:33.841","I2C","R","M","---","100","0x18","3","00 00 00"
"2015-03-12 14:57:33.845","I2C","W","M","--S","100","0x18","1","2D"
"2015-03-12 14:57:33.848","I2C","R","M","---","100","0x18","3","EF EF EF"
"2015-03-12 14:57:33.976","I2C","W","M","--S","100","0x18","1","28"
"2015-03-12 14:57:33.984","I2C","R","M","---","100","0x18","9","E0 E0 E0 E0 E0 E0 E0 E0 E0"
"2015-03-12 14:57:33.993","I2C","W","M","--S","100","0x18","1","29"
"2015-03-12 14:57:34.000","I2C","R","M","---","100","0x18","9","00 00 00 00 00 00 00 00 00"
"2015-03-12 14:57:34.118","I2C","W","M","--S","100","0x18","1","2A"
"2015-03-12 14:57:34.127","I2C","R","M","---","100","0x18","6","90 90 90 90 90 90"
"2015-03-12 14:57:34.135","I2C","W","M","--S","100","0x18","1","2B"
"2015-03-12 14:57:34.144","I2C","R","M","---","100","0x18","6","00 00 00 00 00 00"
"2015-03-12 14:57:34.275","I2C","W","M","--S","100","0x18","1","2C"
"2015-03-12 14:57:34.282","I2C","R","M","---","100","0x18","3","10 10 10"
"2015-03-12 14:57:34.290","I2C","W","M","--S","100","0x18","1","2D"
"2015-03-12 14:57:34.298","I2C","R","M","---","100","0x18","3","EF EF EF"
"2015-03-12 14:57:34.430","I2C","W","M","--S","100","0x18","1","28"
"2015-03-12 14:57:34.437","I2C","R","M","---","100","0x18","9","D0 D0 D0 D0 D0 D0 D0 D0 D0"
"2015-03-12 14:57:34.446","I2C","W","M","--S","100","0x18","1","29"
"2015-03-12 14:57:34.453","I2C","R","M","---","100","0x18","9","00 00 00 00 00 00 00 00 00"
"2015-03-12 14:57:34.585","I2C","W","M","--S","100","0x18","1","2A"
"2015-03-12 14:57:34.594","I2C","R","M","---","100","0x18","6","A0 A0 A0 A0 A0 A0"
"2015-03-12 14:57:34.601","I2C","W","M","--S","100","0x18","1","2B"
"2015-03-12 14:57:34.608","I2C","R","M","---","100","0x18","6","00 00 00 00 00 00"
"2015-03-12 14:57:34.724","I2C","W","M","--S","100","0x18","1","2C"
"2015-03-12 14:57:34.727","I2C","R","M","---","100","0x18","3","60 60 60"
"2015-03-12 14:57:34.732","I2C","W","M","--S","100","0x18","1","2D"
"2015-03-12 14:57:34.736","I2C","R","M","---","100","0x18","3","F0 F0 F0"
"2015-03-12 14:57:34.865","I2C","W","M","--S","100","0x18","1","28"
"2015-03-12 14:57:34.875","I2C","R","M","---","100","0x18","9","40 40 40 40 40 40 40 40 40"
"2015-03-12 14:57:34.882","I2C","W","M","--S","100","0x18","1","29"
"2015-03-12 14:57:34.891","I2C","R","M","---","100","0x18","9","00 00 00 00 00 00 00 00 00"
"2015-03-12 14:57:35.023","I2C","W","M","--S","100","0x18","1","2A"
"2015-03-12 14:57:35.032","I2C","R","M","---","100","0x18","6","80 80 80 80 80 80"
"2015-03-12 14:57:35.042","I2C","W","M","--S","100","0x18","1","2B"
"2015-03-12 14:57:35.049","I2C","R","M","---","100","0x18","6","00 00 00 00 00 00"
"2015-03-12 14:57:35.178","I2C","W","M","--S","100","0x18","1","2C"
"2015-03-12 14:57:35.181","I2C","R","M","---","100","0x18","3","20 20 20"
"2015-03-12 14:57:35.186","I2C","W","M","--S","100","0x18","1","2D"
"2015-03-12 14:57:35.188","I2C","R","M","---","100","0x18","3","EF EF EF"
"2015-03-12 14:57:35.302","I2C","W","M","--S","100","0x18","1","28"
"2015-03-12 14:57:35.312","I2C","R","M","---","100","0x18","9","F0 F0 F0 F0 F0 F0 F0 F0 F0"
"2015-03-12 14:57:35.319","I2C","W","M","--S","100","0x18","1","29"
"2015-03-12 14:57:35.325","I2C","R","M","---","100","0x18","9","00 00 00 00 00 00 00 00 00"
"2015-03-12 14:57:35.443","I2C","W","M","--S","100","0x18","1","2A"
"2015-03-12 14:57:35.450","I2C","R","M","---","100","0x18","6","60 60 60 60 60 60"
"2015-03-12 14:57:35.459","I2C","W","M","--S","100","0x18","1","2B"
"2015-03-12 14:57:35.467","I2C","R","M","---","100","0x18","6","00 00 00 00 00 00"
"2015-03-12 14:57:35.598","I2C","W","M","--S","100","0x18","1","2C"
"2015-03-12 14:57:35.604","I2C","R","M","---","100","0x18","3","D0 D0 D0"
"2015-03-12 14:57:35.608","I2C","W","M","--S","100","0x18","1","2D"
"2015-03-12 14:57:35.614","I2C","R","M","---","100","0x18","3","EF EF EF"
"2015-03-12 14:57:35.740","I2C","W","M","--S","100","0x18","1","28"
"2015-03-12 14:57:35.747","I2C","R","M","---","100","0x18","9","B0 B0 B0 B0 B0 B0 B0 B0 B0"
"2015-03-12 14:57:35.756","I2C","W","M","--S","100","0x18","1","29"
"2015-03-12 14:57:35.765","I2C","R","M","---","100","0x18","9","FF FF FF FF FF FF FF FF FF"
"2015-03-12 14:57:35.898","I2C","W","M","--S","100","0x18","1","2A"
"2015-03-12 14:57:35.901","I2C","R","M","---","100","0x18","6","80 80 80 80 80 80"
"2015-03-12 14:57:35.904","I2C","W","M","--S","100","0x18","1","2B"
"2015-03-12 14:57:35.907","I2C","R","M","---","100","0x18","6","00 00 00 00 00 00"
"2015-03-12 14:57:36.036","I2C","W","M","--S","100","0x18","1","2C"
"2015-03-12 14:57:36.045","I2C","R","M","---","100","0x18","3","10 10 10"
"2015-03-12 14:57:36.055","I2C","W","M","--S","100","0x18","1","2D"
"2015-03-12 14:57:36.062","I2C","R","M","---","100","0x18","3","F0 F0 F0"
"2015-03-12 14:57:36.191","I2C","W","M","--S","100","0x18","1","28"
"2015-03-12 14:57:36.200","I2C","R","M","---","100","0x18","9","20 20 20 20 20 20 20 20 20"
"2015-03-12 14:57:36.210","I2C","W","M","--S","100","0x18","1","29"
"2015-03-12 14:57:36.217","I2C","R","M","---","100","0x18","9","00 00 00 00 00 00 00 00 00"
"2015-03-12 14:57:36.348","I2C","W","M","--S","100","0x18","1","2A"
"2015-03-12 14:57:36.351","I2C","R","M","---","100","0x18","6","30 30 30 30 30 30"
"2015-03-12 14:57:36.355","I2C","W","M","--S","100","0x18","1","2B"
"2015-03-12 14:57:36.358","I2C","R","M","---","100","0x18","6","00 00 00 00 00 00"
"2015-03-12 14:57:36.473","I2C","W","M","--S","100","0x18","1","2C"
"2015-03-12 14:57:36.480","I2C","R","M","---","100","0x18","3","30 30 30"
"2015-03-12 14:57:36.489","I2C","W","M","--S","100","0x18","1","2D"
"2015-03-12 14:57:36.496","I2C","R","M","---","100","0x18","3","EF EF EF"
"2015-03-12 14:57:36.630","I2C","W","M","--S","100","0x18","1","28"
"2015-03-12 14:57:36.638","I2C","R","M","---","100","0x18","9","60 60 60 60 60 60 60 60 60"
"2015-03-12 14:57:36.645","I2C","W","M","--S","100","0x18","1","29"
"2015-03-12 14:57:36.655","I2C","R","M","---","100","0x18","9","00 00 00 00 00 00 00 00 00"
"2015-03-12 14:57:36.785","I2C","W","M","--S","100","0x18","1","2A"
"2015-03-12 14:57:36.792","I2C","R","M","---","100","0x18","6","80 80 80 80 80 80"
"2015-03-12 14:57:36.802","I2C","W","M","--S","100","0x18","1","2B"
"2015-03-12 14:57:36.809","I2C","R","M","---","100","0x18","6","00 00 00 00 00 00"
"2015-03-12 14:57:36.940","I2C","W","M","--S","100","0x18","1","2C"
"2015-03-12 14:57:36.948","I2C","R","M","---","100","0x18","3","20 20 20"
"2015-03-12 14:57:36.957","I2C","W","M","--S","100","0x18","1","2D"
"2015-03-12 14:57:36.966","I2C","R","M","---","100","0x18","3","F0 F0 F0"
"2015-03-12 14:57:37.082","I2C","W","M","--S","100","0x18","1","28"
"2015-03-12 14:57:37.085","I2C","R","M","---","100","0x18","9","C0 C0 C0 C0 C0 C0 C0 C0 C0"
"2015-03-12 14:57:37.091","I2C","W","M","--S","100","0x18","1","29"
"2015-03-12 14:57:37.094","I2C","R","M","---","100","0x18","9","00 00 00 00 00 00 00 00 00"
"2015-03-12 14:57:37.221","I2C","W","M","--S","100","0x18","1","2A"
"2015-03-12 14:57:37.230","I2C","R","M","---","100","0x18","6","80 80 80 80 80 80"
"2015-03-12 14:57:37.237","I2C","W","M","--S","100","0x18","1","2B"
"2015-03-12 14:57:37.244","I2C","R","M","---","100","0x18","6","00 00 00 00 00 00"
"2015-03-12 14:57:37.377","I2C","W","M","--S","100","0x18","1","2C"
"2015-03-12 14:57:37.382","I2C","R","M","---","100","0x18","3","C0 C0 C0"
"2015-03-12 14:57:37.390","I2C","W","M","--S","100","0x18","1","2D"
"2015-03-12 14:57:37.394","I2C","R","M","---","100","0x18","3","F0 F0 F0"
"2015-03-12 14:57:37.516","I2C","W","M","--S","100","0x18","1","28"
"2015-03-12 14:57:37.520","I2C","R","M","---","100","0x18","9","50 50 50 50 50 50 50 50 50"
"2015-03-12 14:57:37.525","I2C","W","M","--S","100","0x18","1","29"
"2015-03-12 14:57:37.529","I2C","R","M","---","100","0x18","9","FF FF FF FF FF FF FF FF FF"
"2015-03-12 14:57:37.644","I2C","W","M","--S","100","0x18","1","2A"
"2015-03-12 14:57:37.654","I2C","R","M","---","100","0x18","6","40 40 40 40 40 40"
"2015-03-12 14:57:37.658","I2C","W","M","--S","100","0x18","1","2B"
"2015-03-12 14:57:37.664","I2C","R","M","---","100","0x18","6","00 00 00 00 00 00"
"2015-03-12 14:57:37.782","I2C","W","M","--S","100","0x18","1","2C"
"2015-03-12 14:57:37.792","I2C","R","M","---","100","0x18","3","30 30 30"
"2015-03-12 14:57:37.801","I2C","W","M","--S","100","0x18","1","2D"
"2015-03-12 14:57:37.809","I2C","R","M","---","100","0x18","3","F0 F0 F0"
"2015-03-12 14:57:37.940","I2C","W","M","--S","100","0x18","1","28"
"2015-03-12 14:57:37.948","I2C","R","M","---","100","0x18","9","00 00 00 00 00 00 00 00 00"
"2015-03-12 14:57:37.957","I2C","W","M","--S","100","0x18","1","29"
"2015-03-12 14:57:37.964","I2C","R","M","---","100","0x18","9","00 00 00 00 00 00 00 00 00"
"2015-03-12 14:57:38.095","I2C","W","M","--S","100","0x18","1","2A"
"2015-03-12 14:57:38.104","I2C","R","M","---","100","0x18","6","80 80 80 80 80 80"
"2015-03-12 14:57:38.111","I2C","W","M","--S","100","0x18","1","2B"
"2015-03-12 14:57:38.118","I2C","R","M","---","100","0x18","6","00 00 00 00 00 00"
"2015-03-12 14:57:38.236","I2C","W","M","--S","100","0x18","1","2C"
"2015-03-12 14:57:38.246","I2C","R","M","---","100","0x18","3","60 60 60"
"2015-03-12 14:57:38.256","I2C","W","M","--S","100","0x18","1","2D"
"2015-03-12 14:57:38.262","I2C","R","M","---","100","0x18","3","EF EF EF"
"2015-03-12 14:57:38.391","I2C","W","M","--S","100","0x18","1","28"
"2015-03-12 14:57:38.394","I2C","R","M","---","100","0x18","9","E0 E0 E0 E0 E0 E0 E0 E0 E0"
"2015-03-12 14:57:38.398","I2C","W","M","--S","100","0x18","1","29"
"2015-03-12 14:57:38.401","I2C","R","M","---","100","0x18","9","FF FF FF FF FF FF FF FF FF"
"2015-03-12 14:57:38.516","I2C","W","M","--S","100","0x18","1","2A"
"2015-03-12 14:57:38.525","I2C","R","M","---","100","0x18","6","E0 E0 E0 E0 E0 E0"
"2015-03-12 14:57:38.535","I2C","W","M","--S","100","0x18","1","2B"
"2015-03-12 14:57:38.542","I2C","R","M","---","100","0x18","6","00 00 00 00 00 00"
"2015-03-12 14:57:38.674","I2C","W","M","--S","100","0x18","1","2C"
"2015-03-12 14:57:38.677","I2C","R","M","---","100","0x18","3","80 80 80"
"2015-03-12 14:57:38.681","I2C","W","M","--S","100","0x18","1","2D"
"2015-03-12 14:57:38.686","I2C","R","M","---","100","0x18","3","EF EF EF"
"2015-03-12 14:57:38.815","I2C","W","M","--S","100","0x18","1","28"
"2015-03-12 14:57:38.818","I2C","R","M","---","100","0x18","9","20 20 20 20 20 20 20 20 20"
"2015-03-12 14:57:38.822","I2C","W","M","--S","100","0x18","1","29"
"2015-03-12 14:57:38.825","I2C","R","M","---","100","0x18","9","00 00 00 00 00 00 00 00 00"
"2015-03-12 14:57:38.953","I2C","W","M","--S","100","0x18","1","2A"
"2015-03-12 14:57:38.960","I2C","R","M","---","100","0x18","6","20 20 20 20 20 20"
"2015-03-12 14:57:38.969","I2C","W","M","--S","100","0x18","1","2B"
"2015-03-12 14:57:38.976","I2C","R","M","---","100","0x18","6","00 00 00 00 00 00"
"2015-03-12 14:57:39.095","I2C","W","M","--S","100","0x18","1","2C"
"2015-03-12 14:57:39.104","I2C","R","M","---","100","0x18","3","60 60 60"
"2015-03-12 14:57:39.111","I2C","W","M","--S","100","0x18","1","2D"
"2015-03-12 14:57:39.119","I2C","R","M","---","100","0x18","3","F0 F0 F0"
"2015-03-12 14:57:39.250","I2C","W","M","--S","100","0x18","1","28"
"2015-03-12 14:57:39.256","I2C","R","M","---","100","0x18","9","30 30 30 30 30 30 30 30 30"
"2015-03-12 14:57:39.262","I2C","W","M","--S","100","0x18","1","29"
"2015-03-12 14:57:39.266","I2C","R","M","---","100","0x18","9","FF FF FF FF FF FF FF FF FF"
"2015-03-12 14:57:39.391","I2C","W","M","--S","100","0x18","1","2A"
"2015-03-12 14:57:39.395","I2C","R","M","---","100","0x18","6","80 80 80 80 80 80"
"2015-03-12 14:57:39.401","I2C","W","M","--S","100","0x18","1","2B"
"2015-03-12 14:57:39.407","I2C","R","M","---","100","0x18","6","00 00 00 00 00 00"
"2015-03-12 14:57:39.532","I2C","W","M","--S","100","0x18","1","2C"
"2015-03-12 14:57:39.536","I2C","R","M","---","100","0x18","3","10 10 10"
"2015-03-12 14:57:39.540","I2C","W","M","--S","100","0x18","1","2D"
"2015-03-12 14:57:39.543","I2C","R","M","---","100","0x18","3","F0 F0 F0"
"2015-03-12 14:57:39.673","I2C","W","M","--S","100","0x18","1","28"
"2015-03-12 14:57:39.680","I2C","R","M","---","100","0x18","9","10 10 10 10 10 10 10 10 10"
"2015-03-12 14:57:39.688","I2C","W","M","--S","100","0x18","1","29"
"2015-03-12 14:57:39.697","I2C","R","M","---","100","0x18","9","00 00 00 00 00 00 00 00 00"
"2015-03-12 14:57:39.829","I2C","W","M","--S","100","0x18","1","2A"
"2015-03-12 14:57:39.838","I2C","R","M","---","100","0x18","6","40 40 40 40 40 40"
"2015-03-12 14:57:39.846","I2C","W","M","--S","100","0x18","1","2B"
"2015-03-12 14:57:39.857","I2C","R","M","---","100","0x18","6","00 00 00 00 00 00"
"2015-03-12 14:57:39.986","I2C","W","M","--S","100","0x18","1","2C"
"2015-03-12 14:57:39.990","I2C","R","M","---","100","0x18","3","D0 D0 D0"
"2015-03-12 14:57:39.993","I2C","W","M","--S","100","0x18","1","2D"
"2015-03-12 14:57:39.996","I2C","R","M","---","100","0x18","3","F0 F0 F0"