cancel
Showing results for 
Search instead for 
Did you mean: 

ST1VAFE3BX vAFE validation against reference signals

bwilkins
Associate II

I'd like to validate ST1VAFE3BX vAFE measurement data against test signals as part of hardware design validation of a custom ECG sensor design based on ST1VAFE3BX.

I intend to input reference signals (differential or single ended) into the two bio channels of the ST1VAFE6BX, capture the raw ADC data, convert to mV, and then compare to the reference signals.

- Conducting this test will help validate the data processing, especially reading and interpreting vAFE data from the FIFO, which is different than if reading from registers.

Are there any documents or guides for how to do such tests?

Currently the vAFE waveforms I capture look correct except for the amplitude. 

1 ACCEPTED SOLUTION

Accepted Solutions
bwilkins
Associate II

Answering my own question here that it's the analog front end circuit prior to the ST1VAFE3BX vAFE input terminals which contribute to the signal attenuation in at least the following two ways:

1. DC blocking capacitors

2. Low-Pass Filter

 

The best way to verify the input signal range and experimental validation of conversion code would be with single-ended inputs test signals feeding directly into the vAFE inputs.

View solution in original post

6 REPLIES 6
Federica Bossi
ST Employee

Hi @bwilkins ,

You can refer to our official examples on Github.

In order to give better visibility on the answered topics, please click on 'Accept as Solution' on the reply which solved your issue or answered your question.

Hi @Federica Bossi ,

 

Thanks for the reply. I have seen the code. I'm after other details such as: description of test setup, input signals, ST1VAFE3BX configuration and expected results.

 

Other information:

- configure device in signal-ended configuration, or differential configuration?

- should I attach a signal generator to the electrodes directly, or via ECG pads?

 

Thank you!

Hi @Federica Bossi , here is an example of an experiment I did.

 

Setup:

- Directly attach signal generator to the ECG electrode pads with alligator clips.

- Signal generator output set to sine wave, 1Hz frequency, amplitude specified in the tests below (2mV to 250mV peak-peak)

- Signal generator configured for "High Z" load, differential signal outputs (to match the ST1VAFE3BX electrode configuration of differential mode).

 

ST1VAFE3BX configuration:

- DIFFERENTIAL MODE, GAIN = 16, Zin = 100MOhm
- vAFE only high-performance mode, output data rage = 800 Hz, bandwidth = 45Hz, using FIFO for data transfer with decimation factor of 2, therefore 400Hz sample rate is obtained from the FIFO.
 
Data collection:
- Read vAFE data from FIFO and convert to 16-bit integer
- Plot 400 samples of the vAFE data, which should give one period of the 1Hz input sine wave

Results:
1. First tests show the range of data that the vAFE spans.
- With 200mVp-p and 250mVp-p amplitude sine wave, the vAFE ADC hits the rails at -32,768 and 32,767 
- Reducing amplitude to 150mVp-p fits within the 65536 range of 16-bit
- Based on these results the 16-bit resolution would be: 150/65536 = 0.002289 mV/LSB = 2.289uV/LSB
 
 vAFE_hit_rails.png
 
2. Additional tests with small amplitude: 2mVp-p, 5mVp-p, 10mVp-p and 20mVp-p
- Below shows 3 cycles (1200 vAFE data points) of the 1Hz sine wave
- Some 60Hz noise can be seen riding on the signal

vAFE_small_signal.png

I am trying to make sense of these results with the information in the datasheet:

 

vAFEparameters.png

 

The data sheet says the input range is +/- 25mV with a gain of 16, but my results show I needed 150/2 or +/-75 mV to span the full range of ADC codes.

 

In another thread it is discussed that:

https://community.st.com/t5/mems-sensors/st1vafe3bx-vafe-sensitivity-and-sample-bit-depth/m-p/807763

the vAFE data when read from the FIFO is 16-bits with resolution 1311 LSB/mV, or 0.0007627 mV/LSB

 

But as shown above I'm getting 0.002289 mV/LSB, which is about 3x the expected value.

 

Assuming the datasheet is correct, it seems my ADC counts are much smaller than I expect.

 

I should be able to input a 50mVp-p sine wave and span the full -32,768 and 32,767 count range of the vAFE ADC.

 

Any thoughts on why I'm seeing such differences?

bwilkins
Associate II

Hi @Federica Bossi 

 

I did another test to confirm the conversion of vAFE data from the FIFO to mV value is correct.

 

Procedure:

1. I duplicated this conversion procedure from ST Github into my code, and added a `LOG()` statement to output the mV values to the console:

st1vafe3bx_fifo_data_t test;
test.ah_bio.raw = (int16_t)fifo_data[j + 1] + (int16_t)fifo_data[j + 2] * 256U;
test.ah_bio.mv = st1vafe3bx_from_lsb_to_mv(test.ah_bio.raw);
LOG_INF("%2i: vAFE: %0.4f", i, test.ah_bio.mv);
which would output values at runtime like this
[00:02:34.119,323] <inf> st1vafe3bx:  1: vAFE: 22.8871
[00:02:34.119,384] <inf> st1vafe3bx:  2: vAFE: 22.8993
[00:02:34.119,445] <inf> st1vafe3bx:  3: vAFE: 22.9153
[00:02:34.119,476] <inf> st1vafe3bx:  4: vAFE: 22.9275

I plot the vAFE mV values.

 

2. Simultaneously I collected the ADC counts (using same method as my previous message, which does not use the console) and converted the results to mV using the scaling factor "1311 LSB/mV". I plot the ADC counts and mV values.

 

The following graph shows all results with 150mVp-p sine waveform applied to the ECG electrodes.

- There is a horizontal offset between the data sets, so please ignore that the individual plots are not time aligned. 

- Left axis = ECG raw (ADC counts) (dark blue)

- Right axis = ECG ST (red) and ECG C2G (mV) (light blue)

vAFE_mV_C2G_ST.png

The ECG ST and ECG C2G show the same vertical range; approx +/- 25mV or 50mVp-p.

From this result I believe the conversion from ADC counts to mV I am doing is correct; my mV values match those exactly from ST.

 

My issue is the input signal to the ECG electrodes was 150mVp-p, and I've found here the data shows it as approx 50mVp-p. Somehow the amplitude is 1/3 of what it should be.

 

Could my circuit be doing this, or can vAFE ADC somehow be reducing the ADC counts in some way when the FIFO is configured based on output data rate (ODR) or bandwidth etc?

 

Thank you for any suggestions.

bwilkins
Associate II

@casadeib Hope you might be able to offer some insight here.

bwilkins
Associate II

Answering my own question here that it's the analog front end circuit prior to the ST1VAFE3BX vAFE input terminals which contribute to the signal attenuation in at least the following two ways:

1. DC blocking capacitors

2. Low-Pass Filter

 

The best way to verify the input signal range and experimental validation of conversion code would be with single-ended inputs test signals feeding directly into the vAFE inputs.