2025-10-24 1:17 AM - last edited on 2025-10-24 2:53 AM by Peter BENSCH
Hi
follow can make offset calibration?Place the object 14cm away .Must be in a dark environment?
follow can make Crosstalk calibration? Place the object 60cm away.
my coverglass like follow,The measured distance seems very short. How can I solve this problem?
follow is my read out data
|
Solved! Go to Solution.
2025-10-27 3:15 AM
Hi
As this discussion is too deep level technical question, let's move to email loop to improve the efficiency and continue communicate the Q&A.
I will initialize an email loop and involve region support team.
Br
Zhiyuan.Han
2025-10-24 2:54 AM
Hi
About offset calibration, normally we suggest do offset calibration at short distance(10cm or 14cm) with white target (88% reflectance) in dark condition.
About xtalk calibration, can you share your ranging data in dark condition? normally we suggest do xtalk at reorted distance start to drop, refer below picture.
And you shared ranging data is under the sun, VL53L0CX ranging ability will decrease with IR noise increase, as sun also contain 940nm noise, so sun light will impact VL53L0CX max ranging ability
you can refer UM for more detail.
Br
Zhiyuan.Han
2025-10-24 7:33 PM
Hi,dear Zhiyuan.Han
follow is offset calibration value:14cm
follow is talkcross calibration 60cm not under sun
2025-10-24 7:57 PM
Hi,dear han
I used continue mode,I set Thresh_Low = 250,Thresh_High=255, I made clibration. find without coverglass is OK
if coverglass , it Trigger interrupt when there are not object before sensror.
2025-10-24 8:22 PM
Dear Han,
This question has puzzled me for a long time.
The interrupt will not be triggered when there is no glass cover in front of the sensor and the background is far away without any obstacles in front of the sensor.
when there are glass cover in front of the sensor, The background distance is too far. Otherwise, the interrupt will be triggered even if there are no obstacles in front of the sensor. the reading is 0 or 8190.
when there are glass cover in front of the sensor,The background distance under 1 meter, all OK。
thank you
2025-10-25 12:43 AM
Hi,
I positioned the glass cover in front of the sensor and performed offset calibration at 14cm. The actual and measured distances are as follows. Where should I perform the CROSSTALK calibration?
2025-10-26 7:50 PM
When there is glass in front of the sensor, if the sensor is not well calibrated, the cover glass will be detected as target, that's why you will receive the abnormal interrupt.
you can check the range status, if status is not 0, then you can discard the data.
and you can also try to play with interrupt, below options are available:
No Interrupt
Level Low (value < thresh_low)
Level High (value > thresh_high)
Out Of Window (value < thresh_low OR value > thresh_high)
Zhiyuan.Han
2025-10-26 7:57 PM
Hi
From the data seems your calibration is not succeeded, or you forgot set the offset value after rebooting, normally after doing offset, if you do range measurement, the reported distance will be around 140mm, but your measurement data is 170mm.
Anyway, from your test result, I suggest you do xtalk between 650~700mm.
Br
Zhiyuan.Han
2025-10-27 1:08 AM
Hi,dear Han
2025-10-27 1:19 AM
Hi,dear Han
every power up ,I must run